(Copied over from email correspondence.)
I believe I remember a conversation in this group about two versus three electrode systems before winter break. It was in the back of my mind for a while so I decided to do some looking into it today to try and wrap my head around it. Using the following measurements we obtained during the Edinburgh training, and some extensive conversations with ChatGPT I was able to get some estimations on how off our voltage measurements may be due to using a two-electrode setup and some general takeaways. Iāve attached a heavily edited ChatGPT summary of what I looked into. I may be naĆÆve in my understanding of all this and you guys already knew all of this but Iād love to hear your thoughts and for some double checking on it.
Measurements used:
6.48MS at 18.8C
6.55 MS at 19.1C
(Iām assuming the MS is mS/cm)
| Percent Output (as shown on software) | Voltage |
|---|---|
| 1% | 2.510 |
| 2% | 2.619 |
| 3% | 2.720 |
| 4% | 2.8 |
| 5% | 2.913 |
| 6% | 2.961 |
| 7% | 2.97 |
Summary:
In a two-electrode electrolysis system, the voltage measured between the anode and cathode significantly overestimates the electrochemical potential that is directly relevant to electrode reactions (and that in the media). This is because the measured voltage includes several contributions that do not correspond to the electrode surface potential, which is the quantity that governs electrochemical reactions such as water splitting.
Voltage Contributions in a Two-Electrode System
The measured voltage can be approximated as:
V_āmeasuredā =V_āanodeā +IR_āsolutionā +V_ācathodeā
Where:
V_āanodeā : Overpotential at the working electrode (anode) interface
IR_āsolutionā : Ohmic voltage drop across the electrolyte between electrodes
V_ācathodeā : Overpotential at the counter electrode (cathode) interface
Only the electrode surface potential (i.e., the potential of an electrode relative to a reference electrode) determines whether a given electrochemical reaction occurs. The solution IR drop and counter-electrode overpotential increase the voltage required from the power supply but do not directly increase the chemical driving force at the electrode surface.
Estimation of Voltage Contributions in the Present System
Solution conductivity: ~6.5 mS/cm
Electrode spacing: ~1.5 cm
Electrode dimensions: 6 mm diameter, 30ā35 mm platinized length
Current: ~12 mA
-
Solution resistance and IR drop
R_āsolutionā =Ļd/Aā128" " Ī©
V_āIRā =IR_āsolutionā ā1.5" V" -
Electrode overpotentials (typical values)
Platinum anode: ~50ā150 mV
Stainless steel cathode: ~300ā800 mV
Combined electrode overpotentials: ~0.4ā0.9 V
Total voltage not directly contributing to electrode surface potential
V_āIRā +V_āoverpotentialsā ā1.9"-" 2.4" V"
Thus, a measured voltage of 2.5V likely corresponds to an actual electrode surface potential of only ~0.1-0.6, depending on current. Importantly, this ātrue electrode potentialā is the potential at the electrodeāsolution interface that governs electrochemical reactions.
Observations
High voltage is measured even at very low % output due to fixed IR drop and electrode overpotentials.
Voltage increases only modestly as output increases, indicating most of the measured voltage is not contributing to increased electrolysis.
Significant electrochemical reactions are expected to begin only once the electrode surface potential crosses practical reaction thresholds, predicted to begin only at ~3ā5% output (cannot remember if this was the case or not, will check when possible).
Relation to Water Splitting Thresholds
The commonly cited 1.6ā2.0 V required for water splitting refers to the electrode surface potential relative to a reference electrode (e.g., SHE or Ag/AgCl), not the total voltage measured between two electrodes.
In a two-electrode system, achieving this surface potential typically requires several volts of total applied voltage, depending on solution resistance and counter-electrode kinetics.
Therefore, observing water electrolysis only at higher measured voltages is fully consistent with electrochemical theory.
Comparison to Three-Electrode Sydow Numbers
The voltage ā2.0 V vs Ag/AgCl refer explicitly to the working electrode surface potential, measured independently of solution resistance and counter-electrode effects.
These values cannot be directly compared to two-electrode voltages (e.g., 2.5ā3.5 V).
In practice, achieving such electrode surface potentials often requires similar or higher total applied voltages than those observed here.
Electric Field Experienced by Cells
Solution voltage drop: ~1.5 V across ~1.5 cm ā ~100 V/m
Voltage across a bacterium (~1 µm): ~100 μV
Typical bacterial membrane potential: ~100ā200 mV
Electroporation threshold: ~10ā¶ā10ā· V/m
Cells are very likely safe from electrical harm
Electrophoretic Effects on Cells
Bacterial surface charge is small
Estimated drift velocity under ~100 V/m: ~1 µm/s
This is negligible compared to diffusion and convective mixing
No significant accumulation of cells at the anode or cathode is expected
Key Takeaways
Two-electrode voltage measurements substantially overestimate the electrochemical potential relevant to reactions.
Most measured voltage is expended overcoming solution resistance and electrode kinetics.
The chemically relevant quantity is the electrode surface potential.
The electrical environment experienced by cells is mild and unlikely to inhibit growth or cause damage.